A method is proposed for measuring the relative contribution of extracolumn and column effects to the detrimental interactions which occur between metal-sensitive solutes and the complete HPLC system. The method involves the substitution of a length of narrow bore silica tubing for the column and measuring the extracolumn contribution, which is subtracted from the total bandspreading measured with a column in place to yield the column contribution. The investigation focussed on HILIC separations, which have been relatively little studied compared with similar effects in RPLC. Metal-solute interactions can lead to tailing peaks and reduced sensitivity or even irreversible adsorption of particularly challenging solutes such as mono-, di- and triphosphorylated nucleotides, which show strong interactions between their phosphate groups and metals. A deactivated HILIC column, treated by a vapour deposition procedure gave generally good results when using high pH (pH 9.0) mobile phases, which suppress the effects of metals. The addition of metal complexing agents such as citrate at low millimolar concentration gave further improvements in peak shape at high pH, and even micromolar concentrations of citrate or medronic acid showed good results. These lower concentrations are more favourable for LC-MS. Addition of the higher concentration of citrate gave acceptable results for the nucleotides even at low pH (pH 3.0). With the standard UHPLC instrument used, loss of efficiency due to metal solute interactions was 25% or less, with most losses due to interactions with the column, although this result will depend on the condition and design of the instrument, which is easily assessed by the proposed procedure.
Read full abstract