We address the misconception that the typical physician dose is higher for CT fluoroscopy (CTF) procedures compared to C-Arm procedures. We compare physician scatter doses using two methods: a literature review of reported doses and a model based on a modified form of the dose area product (DAP). We define thismodified form of DAP, "cumulative absorbed DAP," as the product of the area of the x-ray beam striking the patient, the dose rate per unit area, and the exposure time. The patient entrance dose rate for C-Arm fluoroscopy (0.2mGy/s) was found to be 15 times lower than for CT fluoroscopy (3mGy/s). A typical beam entrance area for C-Arm fluoroscopy reported in the literature was found to be 10.6×10.6cm (112cm2), whereas for CTF was 0.75×32cm (24cm2). The absorbed DAP rate for C-Arm fluoroscopy (22mGy*cm2/s) was found to be 3.3 times lower than for CTF (72mGy*cm2/s). The mean fluoroscopy time for C-Arm procedures (710s) was found to be 21 times higher than for CT fluoroscopy procedures (23s). The cumulative absorbed DAP for C-Arm procedures was found to be 9.4 times higher when compared to CT procedures (1.59 mGy*m2 vs. 0.17 mGy*m2). The higher fluoroscopy time in C-Arm procedures leads to a much lower cumulative DAP (i.e., physician scatter dose) in CTF procedures. This result can inform interventional physicians deciding on whether to perform inter-procedural imaging inside the room as opposed to retreating from the room.