Recently, cavitation on the surface of mechanical heart valves has been studied as a cause of fractures occurring in implanted mechanical heart valves. The cause of cavitation in mechanical heart valve was investigated in both 25-mm Björk-Shiley and 25-mm Medtronic Hall valves. The closing events of these valves in the mitral position were simulated in an electrohydraulic total artificial heart with a stroke volume of 85 ml. The tests were conducted under physiologic pressures at heart rates of 60, 70, 80, and 90 beats/min with cardiac outputs of 4.5, 5.5, 6.4, and 7.5 l/min, respectively. The disk closing behavior was measured by a laser displacement sensor. The closing behaviors were investigated under various atrial and aortic pressures. In both valves, the duration of closing decreased with an increase in the cardiac output. The greater the amount of atrial pressure, the shorter the closing duration of both valves. The maximum closing velocity of the Medtronic Hall monostrut valve ranged from 0.8 to 0.9 m/s, and that of the Björk-Shiley monostrut valve ranged from 0.73 to 0.78 m/s. In both valves, the maximum closing velocities were less than the reported cavitation thresholds. This suggests that there should be no possibility of occurrence of cavitation in an electrohydraulic total artificial heart with mechanical heart valve.