Development of magnetic resonance guided radiation therapy (MRgRT) has sparked interest in evaluating the performance of ion chambers in the presence of magnetic fields. The effect of the field on electron trajectories alters ion chamber response [1] . In most Monte Carlo (MC) simulations, the geometric sensitive volume, often beginning at the edge of chamber’s stem, is used instead of the potentially unknown true collection volume [2] . This work evaluates the sensitivity of chamber response in the presence of magnetic fields to the collection volume used in the MC calculation. The egs_chamber application of the EGSnrc system is used with a recently validated magnetic field transport algorithm [3] to simulate the response of several ion chambers. The chambers are simulated in a PMMA phantom with an incident Co-60 photon beam and magnetic fields between 0 and 2 T, perpendicular to both the incoming photon field and the long axis of the ion chamber. The dose, normalized to 0 T, is calculated in the geometric sensitive volume with either the first 0, 0.5, or 1.0 mm of the volume away from the stem excluded. Increases in chamber response with a maximum of 1.77 ± 0.03% and 3.33 ± 0.03% are observed for a reduction in the length of the collection volume by 0.5 mm and 1.0 mm, respectively. For four chambers, the reduced volumes generally give better agreement with experimental results [4] . Various chamber orientations are under investigation to minimize the effect. This is an important effect that must be addressed to ensure proper calibration of MRgRT machines.