Compare the degree of acute vascular injury caused by a polymer-free, thin-strut drug-eluting stent (DES) to that caused by a bioresorbable polymer, thick-strut DES using optical coherence tomography (OCT). Fifty patients requiring nonurgent PCI were randomized to receive either a thin or a thick-strut DES. OCT was performed before and after stent implantation. OCT-based injury score (IS) after implantation was numerically higher within thick-strut stents 0.32 vs. 0.23, but the difference was NS (P = 0.61). Edge dissections were present in 36% of the patients without differences between groups. Tissue prolapse (TP) area was larger with thin-strut stents (2.26 vs. 1.83 mm2, P = 0.04). Stent expansion and symmetry index were similar between the two platforms (85% vs. 94%, P = 0.08; and 0.82 vs. 0.80, P = 0.25). No differences were observed in total malapposition area (1.85 mm2 in thin-strut stents vs. 1.47 mm2, P = 0.48). Regarding the influence of plaque-type, IS tended to be higher (non-significant) with thick strut DES in fibrocalcific plaques. Stent malapposition area was smaller in fibrous plaques, especially with thin strut stents (P = 0.03). There was no difference in the extent of OCT-based vessel injury associated with thin and thick-strut DES platforms. TP was larger with the thin strut DES, potentially reflecting a deeper stent embedment in the vessel wall.
Read full abstract