The aim of the presented retrospective study was to evaluate the early crestal bone changes around an implant type designed for high primary stability. A total number of 111 implants placed clinically were evaluated regarding insertion torque, bone density, implant stability quotient (ISQ) and early crestal bone loss from standardized digital radiographs. The implants were allocated in two groups: the „regular torque “ group contained all implants that achieved less than 50 Ncm as final insertion torque (n = 63) and the „high torque“ group contained the implants that achieved 50–80 Ncm (n = 48). To avoid possible damage either to the implant´s inner connection or to the bone by application of excessive force, a limit of 80 Ncm was set for all surgeries. All implants underwent submerged healing for three months. ISQ measurements and standardized digital radiographs were taken at day of insertion and at day of second stage surgery. The bone loss was measured on the mesial and distal aspect of the implant. The data evaluation showed the following results: Mean bone loss was 0.27 ± 0.30 mm for the high torque group and 0.24 ± 0.27 mm for the regular torque group. The difference was not statistically significant (p = 0.552). In the two groups, no complications nor implant loss occurred. For the evaluated implant type, there was no significant difference in crestal bone changes and complication rate between high and regular insertion torque in the early healing period.Graphical