For Doppler velocity measurements, Intrinsic Spectral Broadening (ISB) is a broadening of the Doppler spectrum due to the finite spacing between the transducer elements, leading to a range of insonation angles. This effect leads to overestimation in velocity and a decrease in quantification precision. There have been limited attempts to quantify the Intrinsic Spectral Broadening (ISB) of the Aixplorer system under a wide range of imaging conditions. This could be of clinical relevance as it remains unclear whether ultrafast Doppler suffers from the same degree of imprecision as conventional Doppler imaging. In this work, the Aixplorer system ISB was measured using a Doppler string phantom for a range of velocities (10–200 cm/s) and vessel depths (1–4 cm). The results of this analysis showed that, for all depths, the ISB could be modelled using a power law fit with respect to velocity, with a percentage ISB in the range of 3–10 ± 2%. With respect to depth, the ISB did not appear to change significantly, with a percentage ISB in the range of 3–4 ± 2% measured. When utilising ultrafast acquisitions, the percentage ISB was 0.44–0.63 ± 0.07%, affecting an ISB reduction of a factor of 6.5 ± 0.3. These results indicate that the velocity dependence of ISB is strong, with low velocities suffering a high degree of overestimation. Despite the expected result of a strong dependence of ISB on depth, none was observed. Finally, by utilising ultrafast acquisitions, the percentage of ISB can be dramatically reduced.