This work presents a direct and straightforward approach for the determination of trace elements in fish muscle, oyster, and bovine liver via direct solid sample analysis (SS) using high-resolution continuum source graphite furnace atomic absorption spectrometry (HR-CS GF AAS). The preliminary studies revealed the presence of spectral interferences at the analytical line of Ni at 231.096 nm, which could be corrected subtracting the spectrum of SiO and PO from the sample spectra using least-squares background correction. Moreover, all meat samples were proven homogeneous according to the homogeneity factor (H e) (all values were <10 mg½). Pyrolysis (Tp) and atomization (Ta) temperatures were studied and optimized as 800 °C (Tp) and 2500 °C (Ta) for Mn, 700 °C (Tp) and 2600 °C (Ta) for Ni, and 600 °C (Tp) and 2500 °C (Ta) for Rb. Calibration against aqueous standards was proven feasible for Mn determination, whereas Ni and Rb required calibration against solid standards for their quantification. The detection limits achieved were demonstrated adequate for application to food analysis (0.005 μg g−1 for Mn, 0.002 μg g−1 for Ni, and 0.1 μg g−1 for Rb). The developed method was successfully applied for the elemental analysis of fish muscle, oyster, and bovine liver and three certified reference materials, demonstrating good agreement with the certified values and with the reference technique at a 95 % statistical confidence level.