Abstract. Direct analysis in real time mass spectrometry (DART-MS) has recently emerged as a promising approach for measuring semi-volatile organic compounds (SVOCs) on indoor surface films. However, its broader application in indoor environments is limited by low measurement repeatability and no separation of isomers. Herein we developed a sampling suite of indoor surface films for DART-MS analysis, optimized settings of DART to obtain higher analytical performance, and demonstrated the possibility of separating isomeric compounds using tandem mass spectrometry (MS/MS). Two pairs of isomeric phthalate esters, including di(2-ethylhexyl) phthalate (DEHP) and di-n-octyl phthalate (DnOP) and diisobutyl phthalate (DiBP) and di-n-butyl phthalate (DnBP), were used as examples for method optimization and validation. Under optimized conditions, the instrument responses for all four compounds exhibited good linearity (r>0.992) and acceptable repeatability (intraday relative standard deviation (RSD) < 11.0 %). The limits of quantification for the four phthalate esters ranged from 0.042 to 0.24 ng cm−2. The uncertainty in the separation of isomeric components using MS/MS was < 11.4 %, which is acceptable for real sample analysis. To further assess the developed method, we analyzed 10 film samples collected side by side in an occupied office. DnOP was not detected. The RSD among samples was 6.1 % for DEHP, 4.6 % for DnBP, and 10.4 % for DiBP, indicating the overall good repeatability of the collection and measurement method developed. With improved performance, the developed method increases the feasibility of the DART-MS technique for monitoring the dynamics of chemical composition of indoor surface films.