In this study, we compared the performance of an UltraFast LabChip (UL) V280 system for Clostridium difficile detection in stool with that of Xpert C. difficile/Epi and VIDAS CDAB. Among 176 stool specimens, UL V280 detected toxigenic C. difficile in 22 (22/176, 12.5%) with a sensitivity, specificity, positive predictive value, negative predictive value (NPV) of 100.0%, 99.4%, 99.5% and 100.0%, respectively, which were higher than 95.2%, 97.4%, 83.3%, and 99.3% of Xpert C. difficile/Epi (P > 0.05). Notably, the sensitivity and NPV of ULV280 were significantly higher than those of VIDAS CDAB 52.4% (P < 0.001, odds ratio [OR] = 20.0, 95% confidence interval [CI] = 2.26–176.81) and 93.8% (P = 0.002, OR = 10.27, 95% CI = 1.30–81.17). UL V280 turnaround time (35 min) and cost (6.24 Dollars [$]) per specimen were less than those for Xpert C. difficile/Epi (47 min, 59.26 $) and VIDAS CDAB (65 min, 11.70 $). UL V280 possessed an analytical sensitivity limit of 2500 CFU/ml, 95% [CI] = (Ct: 30.76-34.90), and no cross-reactions with other pathogens were found. The study demonstrates that UL V280 based on a microfluidic chip is a rapid, accurate, easy, and cost-effective diagnostic test for toxigenic C. difficile in stool.