PurposeTo determine the feasibility of a prototype needle that enhances vitreous reflux (VR) to control intraocular pressure (IOP) in intravitreal injection (IVI).MethodsWe created an eye model to compare IVI using a standard 30-G needle with four different versions of a 30-G prototype needle with one to four surface grooves that enhanced VR. We injected 50, 70, and 100 µl saline through porcine sclera or 460-µm-thick rubber and measured the peak and 3-second pressure before we extracted the needle and measured the 10-second pressure.Results50-µl injection through sclera with the standard needle resulted in mean (SD) pressure of 58.6 (3.8) mmHg at peak, 52.8 (4.7) mmHg at 3 s, and 39.6 (18.0) mmHg at 10 s. The prototype needle lowered the pressure; four grooves resulted in mean (SD) pressure of 29.4 (5.6) mmHg at peak, 22.0 (3.7) mmHg at 3 s, and 7.2 (6.6) mmHg at 10 s. 70-µl and 100-µl injections through sclera with the standard needle resulted in mean (SD) pressure of 68.8 (3.6) and 86.0 (6.0) mmHg at peak. Similar to 50-µl injection, the prototype needle lowered the pressure for 70-µl and 100-µl injections. At 10 s, we observed varying leakage at the injection site for sclera but not for rubber.ConclusionsThe study provides proof of concept for a needle design for which surface grooves enhance VR and counteract the effect of IVI on IOP. The safety and efficacy of the prototype needle must be studied further in a clinical trial.