Lesion size is reported to become larger as contact force (CF) increases. However, this has not been systematically evaluated in temperature-guided very high-power short-duration (vHPSD) ablation, which was therefore the purpose of this study. Radiofrequency applications (90 W/4 s, temperature-control mode) were performed in excised porcine myocardium with four different CFs of 5, 15, 25, and 35 g using QDOT-MICRO™ catheter. Ten lesions for each combination of settings were created, and lesion metrics and steam-pops were compared. A total of 320 lesions were analyzed. Lesion depth, surface area, and volume were smallest for CF of 5 g than for 15, 25, and 35 g (depth: 2.7 mm vs. 2.9 mm, 3.0 mm, 3.15 mm, p < .01; surface area: 38.4 mm2 vs. 41.8 mm2, 43.3 mm2, 41.5 mm2, p < .05; volume: 98.2 mm3 vs. 133.3 mm3, 129.4 mm3, 126.8 mm3, p < .01 for all pairs of groups compared to CF = 5 g). However, no significant differences were observed between CFs of 15-35 g. Average power was highest for CF of 5 g, followed by 15, 25, and 35 g (83.2 W vs. 82.1 W vs. 77.1 W vs. 66.1 W, p < .01 for all pairs), reflecting the higher incidence of temperature-guided power titration with greater CFs (5 g:8.8% vs. 15 g:52.5% vs. 25 g:77.5% vs. 35 g:91.2%, p < .01 for all pairs except for 25 g vs. 35 g). The incidence of steam-pops did not significantly differ between four groups (5 g:3.8% vs. 15 g:10% vs. 25 g:6.2% vs. 35 g:2.5%, not significant for all pairs). For vHPSD ablation, lesion size does not become large once the CF reaches 15 g, and the risk of steam-pops may be mitigated through power titration even in high CFs.