The inception of the National Breast Screening Programme, following publication of the report Department of Health and Social Security (DHSS), 1986) of a working group under the Chairmanship of Sir Patrick Forrest, led to the construction of a significant number of new mammography rooms including mobile units, with many more being proposed for the next few years. As a result, Radiation Protection Advisers (RPAs) are now being required to give advice on room construction for these units. The Guidance Notes For Health Authorities On Mammographic Equipment Requirements For Breast Cancer Screening published by the DHSS (1987), whilst giving clear guidelines to facilitate standardization on most technical aspects of mammography screening/ assessment centres, do not give any outline of the protection requirements, merely stating in paragraph 3.7.3. that “Radiation protection and operational features must meet the requirements of the local Radiation Protection Adviser.” Such Radiation Protection Advisers will need access to both attenuation data and scatter measurements on which to base their recommendations. This paper presents measurements at mammographic energies of both the attenuation properties of a range of materials that could be used in barrier construction, and phantom scatter measurements. In addition, the application of these results to the situation that will be found in the new screening centre and measurements performed to confirm the conclusions are presented.