The SMART Pass™ (SP) algorithm is a high-pass filter that aims to reduce inappropriate therapy (IT) in subcutaneous internal cardiac defibrillator (S-ICD), but SP can deactivate due to low amplitude sensed R waves or asystole. The association between IT and SP deactivation and management strategies were evaluated, hypothesizing SP deactivation increases the risk of IT and device re-programming, or lead/generator re-positioning could reduce this risk. Retrospective single-centre audit of Emblem™ S-ICD devices implanted 2016 to 2020 utilizing health records and remote monitoring data. Cox regression models evaluated associations between SP deactivation and IT. A total of 348 patients (27 ± 16.6 months follow-up) were studied: 73% primary prevention. Thirty-eight patients (11.8%) received 83 shocks with 27 patients (7.8%) receiving a total of 44 IT. Causes of IT were oversensing (98%) and aberrantly conducted atrial fibrillation (2%). SP deactivation occurred in 32 of 348 patients (9%) and was significantly associated with increased risk of IT (hazard ratio 5.36, 95% CI 2.37-12.13). SP deactivation was due to low amplitude R waves (94%), associated with a higher defibrillation threshold at implant and presence of arrhythmogenic right ventricular cardiomyopathy. No further IT occurred 16 ± 15.5 months after corrective interventions, with changing the sensing vector being successful in 59% of cases. To reduce the risk of IT, the cause of the SP deactivation should be investigated, and appropriate reprogramming, device, or lead modifications made. Utilizing the alert for SP deactivation and electrograms could pro-actively prevent IT.