@Background: modern ICDs use different detection algorithms for the discrimination between ventricular and supraventricular tachycardia. The Morphology Detection algorithm (MD, St. Jude Medical) uses the morphology match (MM) of each intracardiac heart beat with a stored reference ECG-signal of the patient. We compared the intraindividual MM during exercise with rapid atrial pacing (Apace) and developed a new method for the identification of MD responder (MR) and nonresponder (MnR). Methods: in 22 ICD patients (16 m, 614-14 y) standartized bicycle ergometry was performed up to maximum heart rate. MM was recorded for each 10 beat-interval between rest and peak exercise heart rate. Each measurement consisted of 8 consecutive beats and was compared with identical Apace rates. Patients with a MM 91% during Apace, the ROC analysis showed a sensitivity of 93.7% and a specificity of 83.3% for the identification of MR (figure).