The long-term afterglow of scintillators is an important aspect, especially when the light signal from a scintillator is evaluated in the current mode. Scintillators used for radiation detection exhibit an afterglow, which usually comes from multiple components that have different decay times. A high level of afterglow usually has a negative influence on the detection parameters for the energy resolution in spectrometry measurements or X-ray and neutron imaging. The paper deals with the long-term afterglow of some types of scintillators, which is more significant for integral measurement when the current is measured in a photodetector. The range of decay times studied was in the order of tens of seconds to days. Seven types of scintillators were examined: BGO, CaF2(Eu), CdWO4, CsI(Tl), LiI(Eu), NaI(Tl), and plastic scintillator. The scintillators were excited by gamma-ray radiation. After irradiation, the detection unit, along with the scintillator, was moved to a laboratory where the anode current of the photomultiplier tube was measured using a picoammeter for at least a day. The measurements showed that CdWO4 and plastic scintillators have relatively low long-term afterglow signals in comparison to the other scintillators studied.