A new method for palladium and platinum direct determination in environmental samples is proposed by coupling ion chromatography with quadrupole inductively coupled plasma MS. In order to optimise Pd and Pt separation and to minimise interference from matrix in real samples, several anionic and cationic stationary phases have been compared at different mobile phase compositions. In particular, the effect of acidity and of the addition of oxalic acid to the eluent on separation and detection performance has been studied, and the anion-exchange column AG11 turned out to be more suitable. After chromatographic and mass spectrometer parameter optimisation, several potential interferences and the main quality parameters of the method, according to the Eurachem-CITAC recommendations, were evaluated: the detection limit for Pt was 5 ng l −1 while the value for Pd was 230 ng l −1. The method was successfully employed in the determination of platinum group elements in urban road dust and atmospheric particulates and the complete absence of matrix spectral interferences was demonstrated.