We have made the determination of V conform to the requirements of the modern (stabilized temperature) furnace technology where the integrated absorbance (A·s) signals are used to quantitate analyte volatilized into a chamber whose temperature is relatively constant during the period when the analyte peak is measured. Graphite tubes with good pyrolytic coating and fast (maximum power) heating are required. We explored the advantages of specially designed tubes, of a cool-down procedure between the char and atomization steps and of very thin platforms. We found that Mg(NO 3) 2 was advantageous as a matrix modifier. With these conditions we found no problems with several matrices reported by earlier workers to be troublesome, for example HNO 3, phosphate, Fe, Mg and Ca. However metals that form very refractory carbides, such as La, Mo, W and Zr may remain troublesome for V, probably because mixed carbides result which include VC. A group of geological samples was analysed for V. Our recommendation is the use of wall atomization from tubes with good pyrolytic coatings, Mg(NO 3) 2 as a matrix modifier and the cool-down procedure to establish a nearly constant temperature along the tube.