Background Late gadolinium enhancement (LGE) cardiac MRI is the clinical gold standard for non-invasive characterization of myocardial scar [1]. The LGE technique is a contrast enhanced inversion recovery (IR) FLASH sequence, and requires a priori knowledge of the initial inversion time (TI) to produce optimal contrast between healthy myocardium and scar substrate. The TI is typically assessed using a Look-Locker based TI scout sequence. Recent advances [2,3] have enabled successful application of LGE MRI to patients with cardiac implantable devices at our institution. Presence of cardiac devices, such as implantable cardioverter defibrillators (ICD), gives rise to severe offresonance in the myocardium. This produces banding artifacts in standard TI scout images which cannot be used to estimate the initial TI for LGE imaging. This can result in LGE images with poor contrast and delays in patient scanning. In this abstract, we present a modification to the standard TI scout sequence that prominently reduces artifacts in TI scout images. Methods The TI scout sequence is an inversion prepared cine sequence with TrueFISP readout. Severe off-resonance from an ICD produces severe banding artifacts in TrueFISP images. In addition, in IR sequences, severe off-resonance can prevent inversion of spins in the myocardium, giving rise to hyper-intensity artifacts [2,3]. Recently, we demonstrated that a wideband (3.8 kHz) inversion pulse can overcome this off-resonance effect and invert off-resonant spins sufficiently, thereby eliminating hyper-intensity artifacts [2,3]. We modified the TI scout sequence by replacing the conventional inversion pulse (bandwidth: 1.1 kHz) with a wideband inversion pulse (spectral bandwidth: 3.8 kHz). We also replaced the TrueFISP readout with a FLASH readout. The modified sequence was tested on a group of T1 phantoms and one ICD patient. Results Figure 1 shows phantom images of the modified TI scout sequence. The phantom results show that when a flip angle of 5° is used, the modified TI scout sequence can be used to determine the optimal TI quite accurately. At higher flip angles, larger TIs may be underestimated due to signal burn-off. Figure 2 demonstrates the banding artifacts produced when the standard TI scout sequence is used in an ICD patient. The modified TI scout sequence is able to remove these artifacts completely and allow effective determination of the optimal TI. Conclusions We have modified the TI scout sequence by implementing a wideband inversion pulse and FLASH readout. Phantom results and patient studies demonstrate that the modified sequence has none of the image artifacts that occur with the standard sequence, and produces a reliable estimation of inversion time. We expect that the new TI scout sequence will lead to improved application of LGE MRI in patients with implantable cardiac devices. Funding