A fluoroscopy system is a relatively complicated piece of equipment made up of several sub-components. For the operator to achieve good image quality without the overhead of high patient and staff doses, it is essential that all of these components are set up correctly. The automatic exposure rate control (AEC) and image quality are important aspects of a fluoroscopy system. Two principal parameters are used to assess AEC function: air kerma rate at the input phosphor with 1.0 mm copper filtration added and entrance surface dose rate for a 20 cm Perspex phantom. Spatial resolution is measured using the Hüttner type 18 spatial frequency grating. The Leeds TO10 contrast-detail test object is used to measure threshold contrast-detail performance. Median limiting spatial resolution was 1.41 lp.mm-1 for image intensifier fields lying between 18 and 24 cm in diameter. Mean low contrast resolution at 63 kV and 0.40 µGy.s-1 was 2.5%.