Purpose:There are several advantages to utilizing the prone technique for intact breast cancer patients. However, as the topography changes, accompanied by the influence of a supporting breast board and patient treatment couch, the question that arises is to whether there is a concern for contralateral breast dose for intact breast cancer patients being treated with this technique.Methods:An anthropomorphic phantom with breast mounds to duplicate intact breast cancer treatment was planned in prone and supine position. Two tangential beams were executed in the similar manner for as the radiotherapy planning system. For the prone setup, a breast dense foam board was used to support the phantom. A grid of 24 OSL nanodots was placed at 6cm, 4cm, and 2cm apart from the medial border for both prone and supine setups. The phantom was set up using megavoltage imaging and treated as per plan. Additional, a similar study was performed on a patient treated in prone position.Results:Overall, the contralateral breast dose was generally higher for prone setups at all locations especially when close to the medial border. The average mean dose was found to be 1.8%, 2.5% of the prescribed dose for supine respectively prone position. The average of the standard deviation is 1.04%, 1.38% for supine respectively prone position. As for patient treated in prone position average mean dose was found to be 1.165% of the prescribed dose and average of the standard deviation is 9.456%.Conclusion:There is minimal influence of scatter from the breast board. It appears that the volatility of the setup could lead to higher doses than expected from the planning system to the contralateral breast when the patient is in the prone position