The wavelength-integrated absorbance (WIA) and summation of absorbance (∑ lines) of different lines were evaluated to enhance sensitivity and determine B, P and S in medicinal plants by HR-CS FAAS. The lowest LOD for B (0.5mgL−1) and P (13.7mgL−1) was obtained by integration of lines 249.773nm (3pixels) and 247.620nm (5pixels), respectively. The ∑ lines for CS at 257.595nm and 257.958nm furnished LOD=30.5mgL−1, ca. 10% lower than the LOD obtained for the WIA using 257.595nm and 5pixels. Data showed the advantage of WIA over ∑ lines to improve sensitivity for all analytes. Under optimized conditions, calibration curves in the 1.0–100mgL−1 B and 50.0–2000mgL−1 P, S ranges were consistently obtained. Results obtained with the HR-CS FAAS method were in agreement at 98% and 95% confidence level with certified values for B and P, respectively. And results for S were in accordance to non-certified values. Concentrations of B, P, and S in 12 medicinal plants analyzed by the proposed method varied within the 19.4–34.5mgkg−1 B, 719–3910mgkg−1 P and 1469–7653mgkg−1 S ranges.