The volume-clamp method (VCM) is a widely used non-invasive continuous blood pressure (BP) monitoring technique, but determining the volume setpoint (V0) in the open-loop phase can be time-consuming and may cause finger tissue displacement. Additionally, frequent interruptions in BP measurement are required during the closed-loop phase to reset V0. We proposed two new approaches: the proportional value method (PVM) for rapid V0 determination in the open-loop phase and the uninterrupted V0 tracking method (UVTM) in the closed-loop phase. PVM adaptively searches for V0 by analyzing the shape index (Prop) of the plethysmographic (PG) signal under constant cuff pressure, while UVTM introduces a new vascular compliance calculation method in closed-loop phase to track changes in V0 without interrupting BP measurement. We describe the non-invasive continuous BP monitor (NC-BPM) based on these methods. Experimental results show that PVM can accurately determine V0, with a mean arterial pressure error of −1.72 ± 2.80 mmHg when compared to oscillometric method (OSCM). UVTM in NC-BPM produced a mean absolute error (MAE) of 2.44 ± 2.91 mmHg for mean arterial pressure (MAP) in induced BP fluctuation test compared to the Nexfin monitor. In the clinical study with invasive reference, the MAE in measuring MAP by NC-BPM was 4.69 ± 5.39 mmHg. Our results show that PVM and UVTM are highly accurate and reliable, with the potential to integrate effectively into VCM-based BP monitoring devices that could benefit patients, healthcare providers, and researchers.