The sensitivity of the miniature mass spectrometer (mini-MS) is largely restricted by the ion transmission in rough vacuum region. Even though various "in-line" ion transfer devices have improved mini-MS sensitivity, the severe dynamic gas is still weakening the efficiency of ion transmission in this region. Inspired by the "off-axis" ion funnel design in the lab-scale mass spectrometers, a miniature orthogonal injection ion funnel (MO-IF) was developed in this study for the mini-MS with a continuous atmospheric pressure interface. Capable of directing injected ions by 90° and then transport them forward to the downstream skimmer, the MO-IF enabled the separation of ions from the dynamic gas flow jetted out of the inlet capillary. The key factors were optimized for the MO-IF, including the effects of RF amplitude, DC electric fields, and the position of the repeller. Under optimized conditions, the MO-IF minimized the negative effects of dynamic gas and improved the ion transmission efficiency by ∼2-fold in comparison with the in-line injection ion funnel. As a result, a lower limit of detection of 0.5 ng/mL were obtained with good linearity for hypaconitine. Additionally, the MO-IF further decreased the buffer gas pressure in the second vacuum chamber and improved the mass resolution by 1.1-1.5 times at different scan rates.