To test the use of low-cost sub-bandage pressure monitors and pulse oximeters as part of a quality-control measure for graduated compression bandaging in leg ulcer clinics. Twenty-five healthy volunteers (mean age 40 years) providing 50 limbs were bandaged with a four-layer compression bandaging system. The ankle systolic pressure (ASP) was measured using a pulse oximeter (Nellcor NBP-40) before applying the graduated compression bandages. Interface pressure was measured by placing pressure sensors on the skin at three points (2cm above the medial malleolus; the widest part of the calf; and a point midway between them) in the supine and standing positions. The ASP was measured again with the pulse oximeter after the bandage had been applied, and the effect of the bandage on the ASP was recorded. The actual pressure created by the bandage was compared with the required pressure profile. Interface pressures varied with change of position and movement. With the operator blinded to the pressure monitors while applying the bandages, the target pressure of 35-40mmHg at the ankle was achieved in only 36% of limbs ([mean +/- 95% confidence interval]; 32.3 +/- 1.6mmHg [supine]; 38.4 +/- 2.4mmHg [standing position]). With the help of the pressure monitors, the target pressure was achieved in 78% of the limbs. There was no correlation between the pressure monitors and pulse oximeter pressures, demonstrating that the pulse oximeter is not a useful tool for measuring sub-bandage pressures. The results suggest a tool (interface pressure monitors) that is easy to operate should be available as part of quality assurance for treatment, training of care providers and education.