The current global mpox virus (MPXV) outbreak has been declared a Public Health Emergency of International Concern by WHO, with more than 80,000 cases confirmed across multiple continents. Diagnosis is confirmed by PCR of viral DNA from vesicle and other swabs. The aim of this study was to assess commercial RT PCR assays for Orthopoxvirus (OPX) and MPXV for analytical sensitivity, and percent agreements and compare them to primer/probe sets employed at the Victorian Infectious Diseases Reference Laboratory (VIDRL), Centers for Disease Control andPrevention (CDC) and US Army Medical Research Institute of Infectious Diseases (USAMRIID). Limits of detection (LOD), intra-run variability, cross-reactivity and performance on forty clinical samples was assessed on eleven commercial assays and five primer/probe combinations used at VIDRL, CDC and USAMRIID. All assays were able to detect OPX and MPXV (LOD 57 to 14,495 copies/mL) with intra-run coefficients of variation between Cycle thresholds of 0.58 and 3.44, and there was no unexpected cross-reactivity. All assays demonstrated 100% negative percent agreement with clinical samples and all but one yielded 100% positive percent agreement. Variations in LOD between assays may be dependent on the platform used and sample type. Despite the overall comparable performance of the assays assessed, it is important that routine laboratories perform in-house validations before implementing RT PCR for OPX and/or MPXV as reliable and accurate laboratory diagnosis of MPXV and isolation is crucial to containing the spread of this current outbreak and informing public health interventions and response.