The growing importance of monoclonal antibodies and virus particles has led to a pressure for faster size exclusion chromatography. In recent years, numerous small particle columns for size exclusion chromatography of biologicals have been introduced. Small particles are a strategy to reduce analysis time. In the following study, opportunities of small particles in size exclusion chromatography of large biomolecules are investigated. Poppe plots reveal that the lower particle size limit depends on the size of the sample molecule. Hydrodynamic radii of monoclonal antibody monomer, aggregates and H1N1 as well as the diffusion coefficients were determined. Considering this sample compound dependency, kinetic plots referring to the resolution of a distinct compound pair instead of the plate number of a single analyte are more meaningful. Plate times were found to be equivalent with 4 and 2μm particles for a monoclonal antibody aggregate separation at resolutions smaller than 1.8. Quantification of a H1N1 in clarified cell culture can be accomplished with 17μm and 13μm particles at equal plate times at resolutions smaller than 2.5. Virus polydispersity is likely to be affected by run times of several hours at room temperature and shear forces resulting from particles smaller than 10μm. Comparatively high flow rates should be applied in size exclusion chromatography of the 100nm H1N1 virions.