Twenty-four-hour Holter monitoring (24HM) is commonly used to assess cardiac rhythm after surgical therapy of atrial fibrillation (AF). However, this "snapshot" documentation leaves a considerable diagnostic window and only stores short-time cardiac rhythm episodes. To improve accuracy of rhythm surveillance after surgical ablation therapy and to compare continuous heart rhythm surveillance versus 24HM follow-up intraindividually, we evaluated a novel implantable continuous cardiac rhythm monitoring (IMD) device (Reveal XT 9525). Forty-five cardiac surgical patients (male 37, mean age 69.7+/-9.2 years) with a mean preoperative AF duration of 38+/-45 m were treated with either left atrial epicardial high-intensity focus ultrasound ablation (n=33) or endocardial cryothermy (n=12) in case of concomitant mitral valve surgery. Rhythm control readings were derived simultaneously from 24HM and IMD at 3-month intervals with a total recording of 2021 hours for 24HM and 220 766 hours for IMD. Mean follow-up was 8.30+/-3.97 m (range 0 to 12 m). Mean postoperative AF burden (time period spent in AF) as indicated by IMD was 37+/-43%. Sinus rhythm was documented in 53 readings of 24HM, but in only 34 of these instances by the IMD in the time period before 24HM readings (64%, P<0.0001), reflecting a 24HM sensitivity of 0.60 and a negative predictive value of 0.64 for detecting AF recurrence. For "real-life" cardiac rhythm documentation, continuous heart rhythm surveillance instead of any conventional 24HM follow-up strategy is necessary. This is particularly important for further judgment of ablation techniques, devices as well as anticoagulation and antiarrhythmic therapy.