An experimental method of protein detection by non-gel sieving capillary electrophoresis with laser induced fluorescence(NGS-CE-LIF) was established and improved. The method was applied to the analysis of protein complex of lung cancer and compared to capillary electrophoresis with UV detection (CE-UV) and polyacrylamide gel electrophoresis(PAGE). The protein was derived by fluorescein isothiocyanate. The optimized conditions such as 0.05% (w/V) poly(ethylene oxide) (300000) as sieving medium, laboratory buffer TBE (pH 10.0) as electrophoresis buffer, separation voltage of 15 kV, temperature of 15 degrees C and argon ion laser (lambda(ex) = 488 nm, lambda(em) = 520 nm) as detector were employed. Under the conditions, the proteins under the range of 6.5 - 200 kDa were completely separated in 15 min with good separation efficiency. The average theoretical plate number (N) is 9.12 x 10(4)/m, the detection limit is 2.8 x 10(-4) g/L. The results showed that the improved method has the advantages of good separation efficiency, short separation time and the concentration of sieving medium was easy to adjust. The detection rate and efficiency of the method for the protein complex detection of lung cancer are better than those of CE-UV and PAGE.