Histotripsy is a non-thermal focused ultrasound therapy under development to ablate tissue mechanically via bubble cloud activity. Real-time ultrasound imaging is used for treatment guidance. Bubble cloud hyperechogenicity is reduced in deep abdominal targets, making contrast-specific imaging an active area of research. Subharmonic imaging with chirp-coded excitation can improve bubble cloud detection, though the contrast-to-tissue ratio (CTR) was still limited to nearly 6 dB. Nonlinear components of a signal can be delineated with Volterra filtering, including those associated with bubble oscillations. In this study, we tested Volterra filtering as a means to enhance bubble cloud detection. Histotripsy bubble clouds were generated in scattering tissue-mimicking phantoms. Imaging of the bubble clouds was performed with a chirped pulse (1.9-μs duration, 7–12 MHz bandwidth). The scattered signals were processed with a subharmonic matched filter, followed by a tuned second-order Volterra filter. Volterra filtering improved the CTR for bubble cloud detection two-fold relative to matched filtering alone (12.5 dB vs 5.6 dB). Further improvement in bubble contrast was observed for third-order Volterra filtering (CTR of 20.3 dB) but at the cost of underestimating the bubble cloud area. Overall, these findings indicate the utility of Volterra filtering as a means to improve histotripsy image guidance.