A blood sample from a 14-year-old dog was submitted to the veterinary diagnostic laboratory of the University of Milan for marked leukocytosis with atypical cells. A diagnosis of chronic T-cell lymphocytic leukemia (CLL) was made based on blood smear evaluation and flow cytometric phenotyping. A CBC by Sysmex XT-2000iV revealed a moderate normocytic normochromic anemia. Red blood cells counted by optic flow cytometry (RBC-O) resulted in a higher value than using electrical impedance (RBC-I). The relative reticulocyte count based on RNA content and size was 35.3%, while the manual reticulocyte count was <1%. The WBC count of 1,562,680cells/μL was accompanied by a flag. Manual counts for RBC and WBC using the Bürker chamber confirmed the Sysmex impedance results. Finally the manual PCV was lower than HCT by Sysmex. While Sysmex XT can differentiate between RBC and WBC by impedance, even in the face of extreme lymphocytosis due to CLL, RBC-O can be affected by bias, resulting in falsely increased RBC and reticulocyte numbers. Overestimation of RBC-O may be due to incorrect Sysmex classification of leukemic cells or their fragments as reticulocytes. This phenomenon is known as pseudoreticulocytosis and can lead to misinterpretation of regenerative anemia. On the other side PCV can be affected by bias in CLL due to the trapping of RBC in the buffy coat, resulting in a pink hue in the separation area. As HGB concentration is not affected by flow cytometric or other cell-related artifacts it may represent the most reliable variable to assess the degree of anemia in cases of CLL.