The objective of this study was to determine the accuracy of the A&D BP UB-542 wrist device for home blood pressure (BP) measurement according to the International Protocol of the European Society of Hypertension (ESH). Device evaluation was carried out in 33 patients. The mean age was 50.9±10.1 years, the mean systolic BP was 141.6±22.8 mmHg (range 92 : 189), the mean diastolic BP was 89.2±11.4 mmHg (range 62 : 120), the mean arm circumference was 28.8±3.2 cm (range 23-35), and the mean wrist circumference was 17.1±1.4 cm (range 14-19.5). The protocol requirements were followed precisely. The device passed all requirements, fulfilling the standards of the protocol. On average, the device overestimated the systolic BP by 1.8±7.2 mmHg and diastolic BP by 1.6±5.7 mmHg. These data show that the A&D BP UB-542 wrist device met the requirements for validation by the International Protocol and can be recommended for clinical use in the adult population.