This study aimed to elucidate the short-term surgical outcomes and hemodynamics of the Intuity valve compared to the standard bioprosthesis in Japanese patients. Among the 307 consecutive patients who underwent aortic valve replacement (AVR) between February 2019 and March 2021, the Intuity valve was implanted in 95 patients (Intuity group) and a conventional stented bioprosthesis was implanted in 193 patients (conventional group). After propensity score matching, there was no significant difference in in-hospital mortality between the Intuity (n=2, 3%) and conventional groups (n=0, P=0.490). Operation, cardiopulmonary bypass, and aortic cross-clamping times were significantly shorter in the Intuity group. Although the effective orifice area index, trans-prosthetic mean pressure gradient, and peak velocity were similar between the 2 groups at 1 week postoperatively, the Intuity group showed a better mean pressure gradient and peak velocity at 1 year postoperatively. Complete atrioventricular block requiring permanent pacemaker implantation developed in 2 patients (3%) in the Intuity group and none in the conventional group (P=0.476). Mild or greater paravalvular leakage was present in 8 patients (13%) in the Intuity group and 2 patients (3%) in the conventional group (P=0.095). AVR using the Intuity valve in Japanese patients is satisfactory, with a better valve performance and a low incidence of complete atrioventricular block at 1 year postoperatively.