Accurate and rapid measurement of blood β-hydroxybutyrate (β-OHB) concentrations is critical to identify hyperketonemia in sheep. The objective of this study was to test the performance of Freestyle Optium Neo H and TaiDoc-TD4235 point-of-care (POC) devices for measuring blood β-OHB concentrations in sheep. Venous blood samples from 105 sheep were evaluated with Freestyle Optium Neo H and TaiDoc-TD4235 meters. A reference approach was the laboratory measurement of serum BHB concentrations using the Randox D-3 Hydroxybutyrate reagent kit. Data were analyzed using Passing- Bablok regression and Bland-Altman analysis. The diagnostic accuracy was assessed using serum β-OHB concentration at a cut-off value of 0.8 mmol/L. Passing-Bablok regression analysis revealed an intercept of 0.066 (CI95 : 0.018-0.125) and slope of 1.242 (CI95 : 1.136-1.342) for Freestyle Optium Neo H. Intercept and slope values for the TaiDoc-TD4235 meter were 0.625 (CI95 : 0.539-0.724) and 1.265 (CI95 : 1.044-1.497), respectively. The Bland-Altman plot revealed a bias of 0.224 mmol/L and 0.737 mmol/L for Freestyle Optium Neo H and TaiDoc-TD4235 meters, respectively, compared with the reference method. Observed Total Error (TEobs ) was 56.7% for the Freestyle Optium Neo H and 168.7% for the Taidoc-TD4235. The sensitivity and specificity for the Freestyle Optium Neo H were 100% and 80.9%, respectively, at the threshold of 0.8 mmol/L. The Taidoc-TD4235 demonstrated a sensitivity of 93.3% and specificity of 8.9% at the threshold of 0.8 mmol/L. Because of the large TEobs , the Freestyle Optium Neo H and TaiDoc-TD4235 meters should not be used to quantify blood β-OHB concentrations in sheep.