Left bundle branch area pacing (LBBAP) needs confirmation of left conduction system capture assessed by testing of different electrical parameters. Guidelines recommend the use of an electrophysiological recording-system (EP-RS) to guide conduction system pacing (CSP) procedures. However, some experienced centers perform LBBAP procedures without an EP-RS. We aimed to assess whether LBBAP criteria can be measured using the signals provided by the pacing system analyzers (PSA) as a surrogate for the EP-RS to simplify and universalize the technique. This was an observational, prospective, multicenter study assessing the current LBBAP criteria using the PSA compared to the EP-RS during CSP procedures. 108 consecutive patients were included. Baseline QRS duration was 130±29 vs 128±29msec in the EP-RS vs PSA (p=0.7). An initial "W" morphology in V1 was noted in 88% patients with the EP-RS vs 86% patients with the PSA (p=0.7) during unipolar pacing. The QRS duration (122±17 vs 123±19 msec, p=0.7), the R-wave peak time interval in V6 (80±13 vs 79±14 msec (p=0.9) and the V6-V1 inter-peak interval (39±16 vs 38±17 msec (p=0.7) were superimposable in the EP-RS compared to the PSA. Pearson coefficient for the last two criteria were 0.85 (p<0.0001) and 0.94 (p<0.0001), respectively. According to the current criteria, 91.5% of patients received a successful LBBAP implant using the EP-RS. Based on the PSA measurements, 96.6% of these patients accomplished LBBAP criteria. Criteria for LBBAP can be assessed by PSAs with high accuracy. These results provide the basis for the usefulness of the PSA to guide LBBAP procedures.