Determination of trace lead from environmental samples was carried out by solidified floating organic drop microextraction for preconcentration before analysis by flame atomic absorption spectrometry. The determination of lead from environmental samples, even at trace levels, was possible with an ordinary flame atomic absorption spectrometer (FAAS). The lead formed a hydrophobic complex with the anionic surfactant sodium dodecyl benzene sulfonate (SDBS), and this hydrophobic complex was extracted into the 1-dodecanol drop. Parameters affecting analytical performance have been studied. The optimum pH value was 3.5. The optimum sample volume, SDBS concentration, and extraction solvent volume were 75 mL, 50 mM in dodecanol, and 75 µL, respectively. The optimum extraction time was determined as 30 min, the optimum extraction temperature was 45 °C, the optimum mixing speed was 650 rpm and the ideal extraction phase volume was 0.50 mL. Methanol has been shown to be effective when used as a diluent. Using the optimum conditions, the enhancement factor was 106, and the limit of detection (3s) and precision were 3.3 ng/mL and 2.3% (n = 9,15 ng/mL), respectively. The limit of quantification (LOQ) (10s) was 10 ng/mL and the linear working range was 15-100 ng/mL. The accuracy was evaluated by certified reference material analysis. The optimized method was applied to the determination of Pb(II) from environmental water samples. This method follows the principles of green analytical chemistry.