Mechanical thrombectomy is a fundamental intervention for acute ischemic stroke treatment. While conventional techniques are effective, cyclic aspiration (CyA) shows potential for better recanalization rates. We aim to investigate factors affecting CyA and compare them with static aspiration (StA). StA setup consisted of an aspiration pump connected to pressure transducer. CyA was tested with 5 subsequent iterations: single solenoid valve with air plus saline (i1) or saline alone (i2) as aspiration medium; 2 solenoid valves with air plus saline (i3) as aspiration medium; complete air removal and saline feeding (i4); and pressurized saline feeding (i5). To assess the efficacy of clot ingestion, the pressure transducer was replaced with a distal aspiration catheter. Moderately stiff clot analogs (15 mm) were used to investigate the ingestion quantified as clot relative weight loss. Additionally, the aspiration flow rate was assessed for each setup. With CyA i1, the amplitude of the achieved negative pressure waves declined with increasing frequencies but progressively increased with each subsequent iteration, achieving a maximum amplitude of 81 kPa for i5 at 1 Hz. Relative clot weight loss was significantly higher with i5 at 5 Hz than with StA (100% versus 37.8%; P = .05). Aspiration flow rate was lower with CyA than with StA (i5 at 5 Hz: 199.8 mL/min versus StA: 311 mL/min; P < .01). CyA with the appropriate setup may represent an encouraging innovation in mechanical thrombectomy, offering a promising pathway for improving efficacy in clot ingestion and recanalization. The observed benefits warrant confirmation in a clinical setting.