The valve-sparing aortic root replacement (VSARR) procedure was developed to preserve the aortic valve apparatus to replace aneurysmal aortic roots with synthetic grafts and to eliminate associated aortic regurgitation (AR). However, residual post-repair AR is not uncommon and has been found to be associated with recurrent AR and future reoperation. We designed and manufactured a 3D-printed, external adjustable symmetrically extensible (EASE) aortic annuloplasty ring that can symmetrically reduce the aortic annulus diameter via a radial constriction, compliant mechanism. An ex vivo porcine VSARR model with annular dilation and AR was developed (n = 4) and used for hemodynamic, echocardiography, and high-speed videography data collection. After ring annuloplasty repair using the EASE aortic ring, the regurgitant fraction decreased from 23.6 ± 6.9% from the VSARR model to 7.4 ± 5.6% (p = 0.05), which was similar to that measured from baseline with a regurgitant fraction of 10.2 ± 3.9% (p = 0.34). The leaflet coaptation height after annuloplasty repair also significantly increased from that measured in VSARR model (0.4 ± 0.1cm) to 0.9 ± 0.1cm (p = 0.0004), a level similar to that measured in baseline (1.1 ± 0.1cm, p = 0.28). Using an ex vivo VSARR model, the EASE ring successfully reduced AR by reducing the annular diameter and improving leaflet coaptation. With its broad applicability and ease of use, this device has the potential to have a significant impact on patients suffering worldwide from AR due to root aneurysms.