Recently, trans-radial intervention has gained popularity as a common procedure to reduce hemorrhagic complications. However, the cuff-type hemostatic device (TR Band) previously used at our institution required 6 h to achieve hemostasis. Since July 2016, we have been using the VasoSTAT, a new hemostatic device that could achieve hemostasis in 4 h. In a verification study, we found that prolonged activated clotting time (ACT) was related to transient hemorrhage occurrence after the hemostatic procedure. Therefore, we designed a hemostatic protocol based on ACT and evaluated its efficacy. In this retrospective and observational study, 78 and 111 patients used the VasoSTAT and TR Band, respectively, from July 2015 to May 2017. In the VasoSTAT group, the ACTs were significantly lower in the hemostasis success (246 ± 46 s) than in the failure group patients (327 ± 59 s) (P < 0.01). Therefore, we applied the hemostatic protocol to 271 patients from May 2017 to March 2020. The hemostasis success rate was 96% in the post-protocol applied group patients, which was significantly higher than the 82% success rate in the pre-protocol applied group patients (P < 0.01). VasoSTAT resulted in adequate hemostasis in 4 h. Further, ACT was predictive of adequate hemostasis.