A line of passive, length-of-stain dosimeters has been developed for ammonia, amines, carbon dioxide, carbon monoxide, chlorine, hydrogen cyanide, hydrogen chloride, hydrogen sulfide, nitrogen dioxide and sulfur dioxide. These were described in a previous publication. This paper deals with the calibration and accuracy testing of the ammonia (NH3) dosimeter. These dosimeters operate according to Fick's Law of Diffusion. For the NH3 dosimeter, this law can be modified to the form: 15S+S2=KCt, where S=stain length (mm), K is a constant, C=concentration (ppm) and t=time (hours). By experimentally determining K for each lot of tubes, the concentration corresponding to the stain length can readily be caluclated for a given exposure time within the applicable concentration range. The range of the NH3 dosimeter is from about 1 to 280 ppm. The accuracy of these NH3 dosimeters has been validated internally by using a technique described by DuPont. Accuracies of four recent lots were: ±16.3,19.7, 23.4,14.8%. The acc...