ObjectivesWe compared a 3rd generation (gen) and two 4th gen HIV enzyme immunoassays (EIA) to pooled nucleic acid testing (PNAT) for the identification of pre- and early seroconversion acute HIV infection (AHI). Study Design9550 specimens from males >18 year from clinics attended by men who have sex with men were tested by Siemens ADVIA Centaur® HIV 1/O/2 (3rd gen) and HIV Combo (4th gen), as well as by Abbott ARCHITECT® HIV Ag/Ab Combo (4th gen). Third gen non-reactive specimens were also tested by Roche COBAS® Ampliprep/COBAS® TaqMan HIV-1 Test v.2 in pools of 24 samples. Sensitivity and specificity of the three EIAs for AHI detection were compared. Results7348 persons contributed 9435 specimens and had no evidence of HIV infection, 79 (94 specimens) had established HIV infection, 6 (9 specimens) had pre-seroconversion AHI and 9 (12 specimens) had early seroconversion AHI. Pre-seroconversion AHI cases were not detected by 3rd gen EIA, whereas 2/6 (33.3%) were detected by Siemens 4th gen, 4/6 (66.7%) by Abbott 4th gen and 6/6 (100%) by PNAT. All three EIAs and PNAT detected all individuals with early seroconversion AHI. Overall sensitivity/specificity for the EIAs relative to WB or NAT resolved infection status was 93.6%/99.9% for Siemens 3rd gen, 95.7%/99.7% for Siemens 4th gen and 97.9%/99.2% for Abbott 4th gen. ConclusionsWhile both 4th gen EIAs demonstrated improved sensitivity for AHI compared to 3rd gen EIA, PNAT identified more AHI cases than either 4th gen assay. PNAT is likely to remain a useful strategy to identify AHI in high-risk populations.