BackgroundThe diagnosis of myocarditis by CMR requires the use of T2 and T1 weighted imaging, ideally incorporating parametric mapping. Current 2D mapping sequences are acquired sequentially and involve multiple breath-holds resulting in prolonged scan times and anisotropic image resolution. We developed an isotropic free-breathing 3D whole-heart sequence which allows simultaneous T1 and T2 mapping and validated it in patients with suspected acute myocarditis. MethodsEighteen healthy volunteers and 28 patients with suspected myocarditis underwent conventional 2D T1 and T2 mapping with whole heart coverage and 3D joint T1/T2 mapping on a 1.5T scanner. Acquisition time, image quality, and diagnostic performance were compared. Qualitative analysis was performed using a 4-point Likert scale. Bland-Altman plots were used to assess the quantitative agreement between 2D and 3D sequences. ResultsThe 3D T1/T2 sequence was acquired in 8mins 26s under free breathing, whereas 2D T1 and T2 sequences were acquired with breath holds in 11mins 44s (p=0.0001). All 2D images were diagnostic. For 3D images, 89% of T1 and 96% of T2 images were diagnostic with no significant difference in the proportion of diagnostic images for the 3D and 2D T1 (p=0.2482) and T2 maps (p=1.0000). Systematic bias in T1 was noted with biases of 102ms, 115ms, and 152ms for basal-apical segments, with a larger bias for higher T1 values. Good agreement between T2 values for 3D and 2D techniques was found (bias of 1.8ms, 3.9ms, and 3.6ms for basal-apical segments). The sensitivity and specificity of the 3D sequence for diagnosing acute myocarditis was 74% (95% confidence interval [CI] 49-91%) and 83% (36-100%) respectively, with an estimated c-statistic (95% CI) of 0.85 (0.79-0.91) and no statistically significant difference between the 2D and 3D sequences for the detection of acute myocarditis for T1 (p=0.2207) or T2 (p=1.0000). ConclusionFree-breathing whole heart 3D joint T1/T2 mapping was comparable to 2D mapping sequences with respect to diagnostic performance, but with the added advantages of free-breathing, and shorter scan times. Further work is required to address the bias noted at high T1 values, but this did not significantly impact on diagnostic accuracy.