It is critical to measure the maximum voluntary bite force of patients receiving restorative dentistry. A new device known as "BYTE" has been developed indigenously to measure bite force in humans. The purpose of this study is to evaluate the BYTE device's consistency and accuracy in a lab setting. Testing and calibration were done in the laboratory. The calibration machine with load cell pressed the biting part of the device with various forces from 3 N to 444 N in 3 N increments for two to three seconds each. The recorded force value in Newton by the device was noted down. At numerous standard loads, the minimum accuracy error is 0.333 N, while the maximum is 1.667 N. It marginally underestimates the load with an average accuracy error of 0.833 N. The calibration report showed that the BYTE device is precise and reliableand can be used to measure maximum bite force.