Purpose/Objective(s)To dosimetrically quantify the accuracy of the Delta4 (Scandidos, Sweden) phantom for patient-specific quality assurance of volumetric modulated arc therapy (VMAT) patient plans using SmartArc Pinnacle3 treatment planning system (Philips Medical, Fitchburg WI).Materials/MethodsNine (n = 9) SmartArc patient plans were mapped onto the Delta4 phantom and the plan was exported via DICOM RT to the Delta4 software. Treatment plans consisted of 1-3 arcs with the majority using 2 arcs. The Delta4 phantom was calibrated using a Varian 2100 linear accelerator (Varian Medical, Palo Alto, CA). Treatment plans were delivered using either a Varian 2100 with Millennium 120 MLC or NovalisTX. Measurements were compared to calculated doses using the gamma index and profile comparison. Measurement points evaluated were truncated to those within 20% to 500% of the maximum measured dose. The gamma index evaluation criteria varied from 1 mm/1% to 5 mm/5% to evaluate the accuracy of the radiation delivery.ResultsGood agreement between planned and detector measured dose was observed. For a DTA/dose difference tolerance of 3 mm/3% for the gamma index, the percentage of points satisfying this tolerance were 96,1% (90.5%-99.9%). The percentage of points satisfying the gamma index for the selected tolerance criteria decreased when the tolerance was lowered. With a 2 mm/2%, only three patients' plans satisfied the 90% threshold. For 1 mm/1%, no plans satisfied the threshold.ConclusionsThis preliminary study indicates that Delta4 is clinically useful and show good agreement between measured and calculated doses for VMAT patient plans using SmartArc. Purpose/Objective(s)To dosimetrically quantify the accuracy of the Delta4 (Scandidos, Sweden) phantom for patient-specific quality assurance of volumetric modulated arc therapy (VMAT) patient plans using SmartArc Pinnacle3 treatment planning system (Philips Medical, Fitchburg WI). To dosimetrically quantify the accuracy of the Delta4 (Scandidos, Sweden) phantom for patient-specific quality assurance of volumetric modulated arc therapy (VMAT) patient plans using SmartArc Pinnacle3 treatment planning system (Philips Medical, Fitchburg WI). Materials/MethodsNine (n = 9) SmartArc patient plans were mapped onto the Delta4 phantom and the plan was exported via DICOM RT to the Delta4 software. Treatment plans consisted of 1-3 arcs with the majority using 2 arcs. The Delta4 phantom was calibrated using a Varian 2100 linear accelerator (Varian Medical, Palo Alto, CA). Treatment plans were delivered using either a Varian 2100 with Millennium 120 MLC or NovalisTX. Measurements were compared to calculated doses using the gamma index and profile comparison. Measurement points evaluated were truncated to those within 20% to 500% of the maximum measured dose. The gamma index evaluation criteria varied from 1 mm/1% to 5 mm/5% to evaluate the accuracy of the radiation delivery. Nine (n = 9) SmartArc patient plans were mapped onto the Delta4 phantom and the plan was exported via DICOM RT to the Delta4 software. Treatment plans consisted of 1-3 arcs with the majority using 2 arcs. The Delta4 phantom was calibrated using a Varian 2100 linear accelerator (Varian Medical, Palo Alto, CA). Treatment plans were delivered using either a Varian 2100 with Millennium 120 MLC or NovalisTX. Measurements were compared to calculated doses using the gamma index and profile comparison. Measurement points evaluated were truncated to those within 20% to 500% of the maximum measured dose. The gamma index evaluation criteria varied from 1 mm/1% to 5 mm/5% to evaluate the accuracy of the radiation delivery. ResultsGood agreement between planned and detector measured dose was observed. For a DTA/dose difference tolerance of 3 mm/3% for the gamma index, the percentage of points satisfying this tolerance were 96,1% (90.5%-99.9%). The percentage of points satisfying the gamma index for the selected tolerance criteria decreased when the tolerance was lowered. With a 2 mm/2%, only three patients' plans satisfied the 90% threshold. For 1 mm/1%, no plans satisfied the threshold. Good agreement between planned and detector measured dose was observed. For a DTA/dose difference tolerance of 3 mm/3% for the gamma index, the percentage of points satisfying this tolerance were 96,1% (90.5%-99.9%). The percentage of points satisfying the gamma index for the selected tolerance criteria decreased when the tolerance was lowered. With a 2 mm/2%, only three patients' plans satisfied the 90% threshold. For 1 mm/1%, no plans satisfied the threshold. ConclusionsThis preliminary study indicates that Delta4 is clinically useful and show good agreement between measured and calculated doses for VMAT patient plans using SmartArc. This preliminary study indicates that Delta4 is clinically useful and show good agreement between measured and calculated doses for VMAT patient plans using SmartArc.