Photon-counting computed tomography (PCCT) offers better high-resolution and noise performance than energy integrating detector (EID) CT. In this work, we compared both technologies for imaging of the temporal bone and skull base. A clinical PCCT system and 3 clinical EID CT scanners were used to image the American College of Radiology image quality phantom using a clinical imaging protocol with matched CTDI vol (CT dose index-volume) of 25 mGy. Images were used to characterize the image quality of each system across a series of high-resolution reconstruction options. Noise was calculated from the noise power spectrum, whereas resolution was quantified with a bone insert by calculating a task transfer function. Images of an anthropomorphic skull phantom and 2 patient cases were examined for visualization of small anatomical structures. Across measured conditions, PCCT had a comparable or smaller average noise magnitude (120 Hounsfield units [HU]) to the EID systems (144-326 HU). Photon-counting CT also had comparable resolution (task transfer function f25 : 1.60 mm -1 ) to the EID systems (1.34-1.77 mm -1 ). Imaging results supported quantitative findings as PCCT more clearly showed the 12-lp/cm bars from the fourth section of the American College of Radiology phantom and better represented the vestibular aqueduct and oval and round windows when compared with the EID scanners. A clinical PCCT system was able to image the temporal bone and skull base with improved spatial resolution and lower noise than clinical EID CT systems at matched dose.