BackgroundThe evaluation of particle-bound mercury (PBM) exposure is a crucial aspect of assessing the global cycle of mercury (Hg) and its adverse effects on human health and ecosystems. Nevertheless, the precise and reliable measurement of PBM remains a formidable task because of the costly and cumbersome equipment required, as well as the inadequate sensitivities exhibited by current analytical techniques. In this study, we provided a unique and straightforward approach utilising filter fiber-assisted matrix solid-phase dispersion (FF-MSPD) in conjunction with single-drop solution electrode discharge-induced cold vapor generation atomic fluorescence spectrometry (SD-SEGD-CVG-AFS) for the precise quantification of PBM. The PBM contained in a small filter was efficiently extracted with 200 μL of eluent (0.2 % L-cysteine and 4 % HCOOH) by FF-MSPD and subsequently converted to Hg0 using SD-SEGD-CVG, before being subjected to examination using AFS. ResultsThe resulted limit of detection (LOD, 3σ) was 0.17 pg m−3, obtained with a sample volume of 12 m3, which was much higher than that of the techniques published in the literatures. The aforementioned technique was effectively utilised for the detection of mercury in 19 samples of PM2.5 and PM10 which were collected over a span of several months. SigniffcanceContrast to conventional methods, the proposed method offers a range of distinct advantages, including simplified operation, absence of memory effects, enhanced sensitivity, substantial reduction in reagent usage, and decreased secondary pollution. These advantages are particularly valuable for advancing research on the fate, transport, and exposure routes of environmental mercury.