T -weighted turbo-spin-echo (TSE) sequences are a fundamental technique in brain imaging but suffer from field inhomogeneities at ultra-high fields. Several methods have been proposed to mitigate the problem, but were limited so far to nonselective three-dimensional (3D) measurements, making short acquisitions difficult to achieve when targeting very high resolution images, or needed additional calibration procedures, thus complicating their application. Slab-selective excitation pulses were designed for flexible placement utilizing the concept of k -spokes. Phase-coherent refocusing universal pulses were subsequently optimized with the Gradient Ascent Pulse Engineering algorithm and tested in vivo for improved signal homogeneity. Implemented within a 3D variable flip angle TSE sequence, these pulses led to a signal-to-noise ratio (SNR) improvement ranging from 10%to 30% compared to a two-dimensional (2D) T2w TSE sequence employing -shimmed pulses. field inhomogeneities could be mitigated and artifacts from deviations reduced. The concept of universal pulses was successfully applied. We present a pulse design method which provides a set of calibration-free universal pulses (UPs) for slab-selective excitation and phase-coherent refocusing in slab-selective TSE sequences.
Read full abstract