The analysis of mixtures containing monoclonal antibody (mAb) (approximately 150 kDa molecular weight) and sub-unit impurities (approximately 100 kDa) is challenging, even when adopting the latest ultra-high-pressure liquid chromatography (UHPLC) columns (4.6 mm [Formula: see text] 150mm coated hardware, 1.7 [Formula: see text]m 250 BEH[Formula: see text] Surface-modified Particles) and systems (ACQUITY[Formula: see text] UPLC[Formula: see text] I-class Bio Plus). The main issue still encountered is a persistent tail of the mAb peak. Here, the physical origin(s) of such peak tailing in size-exclusion chromatography (SEC) are investigated from both fundamental and practical approaches. Up to five relevant physical origins are analyzed: sample heterogeneity (glycoforms), UHPLC system dispersion, strong residual binding of the mAb to the SEC particles (via hydrophobic and/or electrostatic interactions) and to the stainless steel column/system hardware, slow escape kinetics of the mAb from the SEC particles, and flow heterogeneity caused by the non-ideal slurry packing of SEC columns. Experiments (testing sample heterogeneity, system dispersion, and strong residual interactions) and calculations (predicting the transient absorption/escape kinetics in a single SEC particle and the two-dimensional peak concentration profiles) altogether unambiguously demonstrate that the observed mAb peak tailing is caused primarily by the long-range velocity biases across the SEC column combined with the slow transverse dispersion of mAbs. Therefore, improvement in the resolution between mAb and sub-unit fragment impurities can only be achieved by increasing the column length, e.g., by applying recycling chromatography at acceptable pressures.