A pre-column fluorescence derivatization method is described for the high performance liquid chromatographic determination of N-terminal tyrosine-containing oligopeptides involving methionine-enkephalin and leucine-enkephalin. The peptides are converted into fluorescent derivatives by heating in a weakly alkaline medium (pH 8.5) containing hydroxylamine, cobalt(II) ion and borate. The derivatives are separated on a reversed-phase column (TSKgel ODS-120T) by gradient elution of acetonitrile in a mobile phase containing borate buffer (pH 8.5) and tetra- n-butylammonium chloride, and then determined by fluorimetry. The derivatization provides a single fluorescent product for each N-terminal tyrosine-containing oligopeptide, but does not allow the production of fluorescent derivatives for peptides having no tyrosyl residue at the N-terminal. The method is selective and sensitive; the lower limits of detection for the N-terminal tyrosine-containing oligopeptides tested were 140–310 fmol per 100 μl injected.