Cerebral vascular angiography, or digital subtraction angiography (DSA), is essential for diagnosing neurological conditions but poses radiation risks. This study aims to analyze the impact of examination parameters and patient characteristics on the radiation dose received during DSA to optimize safety and minimize exposure. A retrospective analysis of 251 DSA procedures using the GE Innova IGS 630 dual-plane instrument was conducted. Data on dose area product (DAP) and air kerma (KERMA), along with patient and examination details, were collected. Statistical analyses, including Mann-Whitney, Kruskal-Wallis, and Spearman rank correlation tests, assessed the relationships between variables and radiation dose outcomes. Significant correlations were found between the sides examined (left, right, or both) and DAP (p < 0.0001) and KERMA (p < 0.0001) values, with bilateral studies showing the highest values. The post hoc Dunn tests showed that the 'L + P' group significantly differs from both the right group (p < 0.0001 and the left group (p < 0.0001). There is no significant difference between the 'P' group and the 'L' group (p-value = 0.53). These results suggest that the right and left (both) group have unique KERMA mGy values compared to the other two groups. A strong correlation (rS = 0.87) existed between DAP and KERMA. The number of projections significantly impacted radiation dose (rS = 0.61). Tube parameters (kV and mA) and skull size had low correlations with DAP and KERMA. Optimizing imaging protocols and individualizing parameters can significantly enhance patient safety and diagnostic efficacy while also reducing occupational exposure for medical staff.